How to shunt cal a load cell using SCXI 1520 and Signal Express

I've searched the boards but still cannot find a thorough solution for my question.

How do I condition and shunt cal a load cell using SigEx?

 

When we check out a load cell from our calibration lab it comes with a shunt "value" for a range of resistances (50k, 100k, 200K ohm is typical). The load cell I'm working with is a 100kip load cell which has a value of 61359lbs @ 100k ohm.

 

With our other signal conditioners, with the shunt applied we adjust the voltage gain until the output reads whatever the shunt value is (in my case it's 61359lbs).

SCXI- 1000 Chassis w/ 1346 adapter
PCI 6281 DAQ card
SCXI- 1520 Bridge Board w/ 1314 Terminal Block (x2)
SCXI- 1180 Feedthrough Panel w/ 1302 Block
Signal Express 2014.
Win7 Enterprise
0 Kudos
Message 1 of 4
(6,628 Views)

Hi, OKors.

 

When you create a DAQmx Strain Task, you can navigate to the Configuration tab » Device tab and select the Calibration button; here, you can perform shunt calibration.  To condition your load cell, you can create a Custom Scale to perform any conversion from the prescaled units measured by a channel to the scaled units associated with your measurement and then apply it to the Task (using the drop-down menu Custom Scaling on the Configuration tab).  Refer to page 121 of the SCXI-1520 manual for more information regarding calibrating.


I hope you're having a great day!

Regards,

Sara Lewandroski
Applications Engineer | National Instruments
0 Kudos
Message 2 of 4
(6,614 Views)

Right, but when I use the Strain Task setup it want's to know the gage factor, which for these load cells we do not have. We calibrate all our load cells in-house, so as I mentioned all we get back from the cal lab is a sticker which gives an output value in force (lbs) when 100k shunt is applied.

 

 

SCXI- 1000 Chassis w/ 1346 adapter
PCI 6281 DAQ card
SCXI- 1520 Bridge Board w/ 1314 Terminal Block (x2)
SCXI- 1180 Feedthrough Panel w/ 1302 Block
Signal Express 2014.
Win7 Enterprise
0 Kudos
Message 3 of 4
(6,611 Views)
0 Kudos
Message 4 of 4