Are you using DAQmx or traditional DAQ device drivers??
01-22-2007 01:26 AM
01-22-2007 12:48 PM
01-22-2007 11:47 PM
No, DAQmx is superior and I strongly recommend you used DAQmx
Elijah, I guess I understood the problem that Sunder is facing, coz even I had experienced it 2 years back
What happens is if you scan only the channels to which sensors/strain gauges are connected, all is fine and you get good strain readings.
But if you happen to include an unconnected channel into the scan list, all the subsequent channels after the open channel will show saturated readings.
Right, Surender??
The problem we found, lies in the settling time of the E series DAQ card's ADC. Even we were using PXI 6071 E
Once the ADC saturates on reading an open channel, it cannot settle fast enough to read the next channel.Hence, the next channel also reads saturated value, instead of the correct value.
This is irrespective of the sampling rate because, this happened even when we did continuous sampling at low scan rates of 10 to 100 samples /sec
We found 2 solutions to this problem
Hope this helps,
Regards,