Signal Conditioning
IEPE Triaxial Accelerometer Issues with NI 9234 Module
Hello NI Community,
I am attempting to connect an Omega ACC 301A triaxial accelerometer with an NI 9234 DAQ module. This accelerometer uses a four pin connector, so I am using a 4 pin connector to 3 BNC terminal cable to connect to the 9234 module. On the spec sheet for the Omega ACC 301A, it states an excitation current of 2 - 6mA @ 15 to 30 VDC, which is in the range of my NI 9234 card and an amplitude range of +/-5 V, which is also in the range of the NI 9234.
In the DAQ assistant, I have all constants set per the accelerometer spec sheet, but when I run the VI, the X axis does not respond to any motion, and the Y and Z axes seem to drift when the accelerometer is motionless. I have attached my VI to this post as well as a screenshot showing a real time plot of data. In the screenshot, the triaxial accelerometer is motionless yet has an almost sinusoidal curve present between -.1 Gs to 0 Gs.
Does anyone know what might be causing this drift on the accelerometers voltage output? Also any reason why the X axis wouldn't respond? This is a brand new accelerometer that was calibrated within the last month.
I have used this same NI 9234 card with other IEPE triaxial accelerometers with no issues. Any advice would be greatly appreciated!
Chris
(5,535 Views)
Hi cj,
Everything looks correctly configured in your DAQ Assistant in LabVIEW. You could try using a scope on the BNC of the X-axis measurements to verify that the accelerometer is actually outputting information. You said it was recently calibrated but it wouldn't hurt to double check. Do the Y and Z axes show drift while the accelerometer is in motion?
Paul C
(5,509 Views)
Hey Cooking_Up_Code,
Unfortunately I do not have a scope here where the testing is being conducted. When I get back to my office, I will be able to test it, but unfortunately I am trying to do field testing this week. This accelerometer was purchased last month, so I highly doubt there is an issue with the accelerometer itself. I honestly think there may be issues with my 4 pin to BNC connector cable. Sadly I only have one cable on me right now, so I can't debug to see if this is the root of the issue.
What is strange is the Y and Z axes seem to react correctly and provide accurate G-loading data, but there seems to be a very low frequency drift where the voltage seems to oscillate. This drift is present even when the accelerometer is in motion. Not sure what would cause a voltage drift over time. Would this be caused by the excitation current oscillating?
Chris
(5,501 Views)
It does sound like it could potentially be an issue with a connection, so double checking it wouldn't hurt. Looking at your DAQ Assistant configuration, you are providing excitation on every axis, is this what your sensor requires? It may be that your Y and Z axes are being influenced by an unecessary excitation current.
Paul C
(5,474 Views)