Signal Conditioning
Level Sensor Output Signal: 240-30Ω, Is it possible to measure with PXIe 4300?
I'm going to buy NI PXIe 4300 with TB4300C for measuring current from many sensors.
One of the sensors has the level sensor that has 240-30 Ω output. I don't want to buy the DMM module.
Is it possible to use PXIe 4300 with TB4300C or TB4300?
Specifications:
Model: 878640
Material: Stainless Steel
Length: 550mm
Protection Rank: IP67
Output Signal: 240-30ohms
Resolution Range: 10-40mm
Operational Temperature: -40°F ~ +185°F (-40℃ ~ +85℃)
(4,217 Views)
Accepted by topic author JJ@TH
At least a link to the datasheet of that sensor would be helpful 😉
What accuracy is needed?
IF that sensor resistor can carry ~20mA current, the easy way would be to add a constant 1.5V supply and use your current measurement. The current sensing resistor of the 4300C is 50Ohm (as far as I found the rigth spec) , so sensor and card input in series to a 1.5V source will give you a current range of 20mA (1.5V/(25Ohm+50Ohm) ) to ~5mA (1.5/(250+50)) .
Senorvalues with a little extention 😉
If you can 'steal' 3.3V- 5V (30mA .. again some buffer)  somewhere in your PXI chassis,  a low drop voltage regulator to 1.5V will do the job. If more accuracy is needed and you are not shure about the stability of your 1.5V source: Measure it 🙂  another current channel and a stable 25Ohm Resistor (4 times a 100Ohm?) and a source supplying 40mA (+buffer) .. voila 😄
An other (more common) method: Use 'any' available AC or DC voltage source and a known reference resistor and spend two voltage input channels. Build a voltage divider with the two resistors (known and sensor) and measure the voltage across them.
If the voltage source is known, only one voltage measurement is needed.
Greetings from Germany
Henrik
LV since v3.1
(4,200 Views)
After a closer look into the 4300C spec .. input impedance is 92 Ohm,
so the supply voltage should be ~2,3 V.
Henrik
LV since v3.1
(4,196 Views)
Thank you for your suggestion Henrik.
Here are the details in the pdf I got from the website
https://goo.gl/9fPjnL
I don't need high accuracy +- 10 to 20 Ω is fine.
(4,191 Views)
Output .. :WTF is Standard American  240-33 ohms ??  😄
Migth work as  suggested with a 2.3V supply
The very simple version is a 5V supply and  a 150 Ohm , your sensor and the 20mA current input as a serial chain.
Still should show the discrete sensor values ...
The measured current is I= <Your voltage supply voltage> / (92 + <Sensor resistance>+<additional series resistor> )
Henrik
LV since v3.1
(4,170 Views)
I think I will use voltage divider.
Thank you.
(4,155 Views)
