Signal Conditioning
Measurement of CJC for 1 of 3 SCC-TC modules
There are multiple SCC modules (> 9) being monitored of various signal types. Is it possible to read the cjc on a t/c/ module along with the other signals in the SC-2345 box
(3,484 Views)
Measuring CJC from TC module on SCC-2345
Hey Slapman,
The SCC-2345 routes the signals like this:
Thermocouple = channel X
CJC = channel X+8
TC module in J1:: Temp reading from channel 0 of DAQ card and CJC from channel 8 of DAQ card.
Also, if you create a DAQmxTask, you can select "built in" for the CJC value and it will automatically be taken into account by the task.
Hope that helped.
Brett
(3,484 Views)
Thanks for the reply Brett"
I agree and understand your response for the t/c measurement
One aspect that I was looking at was the ability to monitor the following configuration effectively:
SCC t/c module in J1 and use J8 for CJC
SCC 1V analog module in J2
SCC 5V analog module in J9
Would this work by acquiring a config of ch 0 for t/c, ch 8 for cjc, ch 1 for V, and ch 9 for V?
Or does it have to go ch's 0,1,8,9
(3,484 Views)