07-14-2008 07:54 AM
07-15-2008
05:08 PM
- last edited on
08-05-2025
02:32 PM
by
Content Cleaner
Hi balaclava,
Since the SC-2043-SG is an older signal conditioning board is does not support programmatic control of its excitation or settings. This is why there is no internal excitation available to select in your DAQmx task. These setting must be configured on the SC-2043-SG using the on-board jumpers.
You will want to set the jumpers for internal power and excitation and set for full bridge gauges. More information on all of the jumper settings can be found in the SC-2043-SG User Manual. Also page 30 of the manual shows how to connect a full bridge sensor.
Once you have the SC-2043-SG configured correctly it then does the bridge signal conditioning and provides the DAQ card with a voltage signal referenced to AI Sense. You can measure this signal using a normal analog input voltage task with the terminal configuration set to NRSE or RSE. It does not provide the DAQ card with a differential signal.
Please let me know if you have any questions and take care.
07-17-2008 02:08 PM