Signal Conditioning
NI 9225_Low Voltage Measurement
I'd like to learn the sensitivity range of NI 9225 (with the 300 V measurement range) when the low voltage signals (2-3 V) are applied as input signals.
*  Can the NI 9225 be used for the low voltage measurement ?
* Can anybody help us with the accuracy of the NI 9225 module especially in low voltage measurement?
Thanks in advance.
(2,251 Views)
I am going to cite some sections from the user manual to help you,
  • Resolution of ADC (24-bit sigma delta) = 50.66uV/LSB
  • Accuracy specs.
Considering the worst-case uncalibrated max, for an input signal of 2V, accuracy will be 2V +/- ( 1.6% of 2V + 0.66% of 425V)
2V +/- (0.032V + 2.805V)
The issue will be that the offset error will be larger than your actual measurement and similarly, you can calculate other conditions.
Santhosh
Soliton Technologies
(2,211 Views)
Thanks for your assistance.
(2,200 Views)
