I recently purchased the NI-USB-6210 to replace my old voltmeter, Keithley 197A. After installation and initial testing, I found that the baseline noise was significant increased.
For the same setting, my old Keithley 197A meter measured 5.0000+/-0.0001 V. However, the new NI-USB-6210 measured 5.00+/-0.01V. The noise is 100 times higher.
At a lower voltage, it's 0.500+/-0.002V VS 0.5+/-0.4V. The noise is too high. According to the spec of 6210, it measures 10 V in 16 bit, and this translates to 0.1 mV. So I suspect the card may be defective.
Does anyone have the same issue? Would this be defective or could it be some settings to increase the resolution?
(5,910 Views)
Well, both instruments use different approaches to measure a voltage...
Usually the most simple trick is avaraging with a windowlength matching (a multiple of) your line frequency. The 197A use filter too 😉
6210 SPEC page 4  tell something about the noise level ...
And if you use multiple channels have a look at the settling error figures on page 2 (note the log(!) scale on the y axis)
I have a 197A in my home lab rack and the only intrument that migth replace it one day would be a HP3458 😄
Greetings from Germany
Henrik
LV since v3.1
(5,902 Views)