Have you considered replacing the 6052E and SCXI-1140 with a single S-series board (such as the PCI-6143) in order to eliminate the phase delay? The catch to this configuration is that you would need to be in parallel mode, so you would need a 6143 for every SCXI-1141. The SCXI-1140 is going to introduce quite a bit of additional error into your measurements that could be eliminated by sending the signal straight to a simultaneous DAQ board. I would recommend looking at the amplitude of your noise and comparing it to the specs of both modules (keep in mind that the noise from the 1141 will be amplified through the 1140 depending on how you have the jumpers set) - this will tell you if your noise levels match the specifications.
Regards,
Logan K
Message Edited by Logan K on 10-13-2006 08:51 AM