05-29-2009 03:04 PM - edited 05-29-2009 03:06 PM
I am running a PXI 1031 chassis with a PXI 6281 multifunction DAQ card conneted to a SCXI 1001 signal conditioning Chassis containing 12 SCXI-1520 modules and SCXI-1314 terminal blocks. The comunicating module (SCXI 1349) is located on module 1. When attempting to read an analog input on ai1 on module 1 it changes the input from all other channels for the entire SCXI 1001 chassis. Changes in the anolog signal on module 1-ai1 are seen in the ouput from all other channels. I have swaped out the SCXI-1520 and terminal block in module 1, with replacements that have proven to be working however the problem persists.
If the communicating module is changed to module 2 on the SCXI-1001 chassis the problem disappears and all the reading appear as the should.
My thinking is that this would be a problem with the backplane of the SCXI chassis.
Thanks
Murray
06-01-2009 06:34 PM
This sounds like a possible hardware malfunction. Here are a couple other suggestions:
- In addition to slot 2, try the 1349 in a couple other slots in the SCXI chassis.
- Test multiple AI channels to see if all are affected, or only a subset of channels.
For further help with troubleshooting and setting up a repair visit www.ni.com/ask where you can call in and speak to an Applications Engineer.