Logan, thank you for the response.
I am having some difficulty selling my Boss with the explanation of the PCI card settling time being "suddenly" the issue.
4 of the 5 test stations with the same clock speed, sampling rate, and software application do not have this issue.
Is there another possible cause for this within the SCXI-1102 hardware, IE: if it was damaged by the computer motherboard capacitor failure? Could all of the SCXI-1102 amplifiers be overloaded or damaged? Could that pulse have damaged the module circuits that convert mV to V?
Thanks Scottyd
09-07-2006 09:05 AM
Scott,
I hope you have your problem solved by now. I am sorry for weighing in on this so late. I was searching the discussion boards and ran across this posting.
I have experienced this when reading ground-referenced thermocouples. In this case, I was using the scxi 1303 terminal block. To resolve this I substituted a 10M-ohm resistor for the factory shipped 10-ohm resistor in the Bias Resistor network. The 10-ohm is only good for floating signals; most thermocouple I use will connect to a grounded test unit.
09-07-2006 01:03 PM
We used and are using 10 M ohm resister packs.
We also switched packs from another TBX1303, just to see if there was a difference, there was none.
it is still an open isssue.
09-08-2006 03:51 PM
09-11-2006 07:10 AM
PCI-6030 board (s/n 11E1F7E)?
old PCI-MIO-16XE-10 boards (s/n AD9C97)
loaner PCI-MIO-16XE-50 boards (s/n 1058A9A).
SR# 7115230.
Their old 16XE-10 failed to work, so they quickly bought the 6030E for short-term replacement so they could keep running while they RMA'd the failed one. It turns out that the 6030 had some other measurement offset issues that only showed up with their SCXI systems and after being RMA'd, so did the repaired 16XE-10. The only board we could get to work in the system was our loaner, PCI-MIO-16XE-50.
This was the guy working on the solution.
Justin Mintzer
Applications Engineer
National Instruments
this person may be able to help you with the rev level issue, the boards are at NI.
Scott Diedrich
Manitowoc Ice Inc.
09-12-2006 09:31 PM
Hello Scott,
Thank you for the information. I looked into the SR to get complete details on the case. I noticed that you had a few working systems while just one exhibited the voltage offset. Did you ever rule out the possibility of there being a problem with the SCXI system and cabling by switching only the SCXI system(and cabling) between a working and non-working system?
You mention that this is still an open issue and I want to make sure you have a long-term solution. Please let us know if there is anything we can do .
Regards,
Jennifer O.
09-13-2006 07:10 AM
Yes we switched TBX1303'S, cable's, modules, connections at SCXI, PCI slots..etc Same O BS
even whole chassis, Same stuff
two different systems had issues, #4 #5
5 test stands running fine, two systems had smoked cap on pc's video cards, then problem appeared.
other 3 still OK
change to the different style board, OK
change back "repaired' board, Bad
Install "new" board, Bad
Intall different (borrowed board), OK
?????
?????
all systems running now with 2 different boards in #4 and #5. had to change software to reflect sampling rate of the different boards to make them run within their spec.
Scottyd