Signal Conditioning
SCXI-1112 Calibration Details
I just received a few SCXI-1112 cards for which I also ordered the detailed calibration.  Some of the information on the detailed calibration report don't appear, at first glance, to match up with the specification on the SCXI-1112 manual.  I'm hoping somebody here can fill in the gaps for me.
First, is there a calibration procedure available for the SCXI-1112 which might answer my questions without a lot of digging?
Now to the details:
According to the detailed calibration report (NI had esz Elektronik-Service GmbH out of Germany perfom the calibration), the calibration lab applied a +90 mV input and a -90 mV input to each channel, and verified that the output from the SCXI-1112 was within ±48 μV of the applied signal.  I'm just not sure where this ±48 μV limit comes from.
The specifications section from the SCXI-1112 manual lists the following:
Nonlinearity: 0.005% of FSR = 10 μV assuming a 200 mV (±100 mV) full scale range.
Offset error after calibration: 20 μV.
Gain error after calibration: 0.005% of a 90 mV reading = 4.5 μV.
If I simply add up these numbers, the output from an SCXI-1112 which meets specifications should be 90 mV ±34.5 μV  (not ±48 μV) with a 90 mV signal applied.  What am I missing here?
Mark Moss
Electrical Validation Engineer
GHSP
(6,328 Views)
Hello MarkMoss,
Thanks for your post and questions about calibration.
You are right in your calculations for the accuracy of the SCXI-1112 and that it should be somewhere in the range of  ±35 μV given that the input is 90mV. The value on your callibarion report also takes into account the system noise of the SCXI-1112, which the manual states is 5 μV of the signal input. There is also an off set of the DMM used to callibrate the 1112 as well. Even though the equipment that is used to calibrate your SCXI-1112 is callibrated there is also some error associated with it. So this could make up the extra 8 μV after taking into account the system noise. Let me know if this address your concerns or questions about the callibration of your SCXI-1112. Have a great day Mark!
Cheers!
Corby_B