I have thermocouples hooked into a SCXI-1122 module. Using virtual channels in MAX, I am able to see all the voltages correctly. Also, by specifying each individual thermocouple in LabView (e.g. ob0! sc1! md4! 1) I am able to see the voltages correctly. When I read more than one thermocouple using the syntax (ob0! sc1! md4! 1,2) I am able to see the correct voltage for only the first thermocouple in the list. I would like to use bulk addressing to read all 16 thermocouples at once (ob0! sc1! md4! 0:15) but it does not show any voltages correctly. I suspect this is a simple configuration problem. Any insight?
(3,129 Views)
Hello;
You are seeing incorrect values probably due to a too low interchannel delay. If you don't specify the interchannel delay of your application, NI-DAQ will set the lowest value allowed, and sometimes that value is too low for the application.
You can change the interchannel delay value at AI Config.vi.
Hope this helps.