You stated that you have an SCXI-1121 that you are trying to read 10-14 V signals off of, but with a resolution of at least 1 mV.
I took a look at the DataSheet for the SCXI-1121 and found that it looks like you are going to be lucky. The gain setting you need is +/-25 V and the accuracy at that range is 728 uV. So it looks like you can measure everything without any additional conditioning and still get the resolution that you need.
Best of luck with this,
07-28-2005 03:55 PM
Otis- Thank you for the feedback, but what is the best way to achieve the overall gain of 0.22 (as depicted in the specifications table for the SCXI 1121). On reviewing the manual, it appears that you can easily set the gain in the SCXI 1121 to "1". This gives an allowable input voltage range of +- 5V.
If I were then to set the 1327 attenuator to 1:1, the 10-14V input would exceed the +- 5V that is allowable on the 1121. Could you please clarify how I should achieve the desired gain value to achieve the +- 25V range?