I’m hoping someone can help me confirm a necessary hardware change with our current SCXI unit.
Here's our current hardware setup:
I was trying acquire a +/- 10v signal through the SCXI- 1180 Feedthrough Panel and have it recognized by the DAQ board. It appears that the 1349 Cable Adapter originally purchased is not the correct one for what we need. Based on this Knowledge Base Article the 1349 adapter does not free up the necessary channels needed to acquire an external analog signal. From what I understand we need a SCXI- 1346 adapter?
I’d like to confirm that this is what we need before the purchase is made and was hoping someone could help verify this for me.
What we’re looking to do is acquire typical strain data using the above mentioned bridge board and terminal block, but also read two channels of signal coming from our hydraulic controller (MTS-407). The controller outputs are a simple +/- 10v signal (LVDT and LoadCell signal, pre-conditioned and calibrated at the controller). My understanding is that we could use the differential amplifier inputs of the feedthrough panel to read this signal- but again our main issue seems to be in our current 1349 Cable Adapter not freeing up the necessary channels on the DAQ board to read the feedthrough panel in the first place.
So please confirm, if I understand this correctly we need to purchase the SCXI- 1346 adapter to read external signals from the Feedthrough panel?
Thank you for your time,
12-11-2007 11:01 AM
I don't believe the 1349 is the issue. The 1346 is intended when you have multiple SCXI chassis -- you indicate a single chassis so the 1349 should be correct.
In describing your problem, you don't identify the DAQ device or the channel of the DAQ card you are using to interface the SCXI chassis and to acquire the signal of the hydraulic controller. Keep in mind, when you connect the SCXI chassis to the DAQ card, channel 0 is typically dedicated to reading the SCXI modules. This is described in the SCXI Getting Started manual, pg 1-17, 1-18. If the hydraulic controller is connected to chn0, the value won't be read.
I would first verify the channel you've connected for the hydraulic controller. If it is channel 0, connect to a different channel. Use MAX to configure the channel and verify the signal. If you still have conflicts, remove the 1520 from the chassis and check for the signal from the hydraulic controller on the appropriate channel.
If you still have problems, identify the chn you are connecting for the hydraulic controller, what you tried, and post your results. Please include the model of DAQ card.
12-11-2007 12:05 PM
12-12-2007 09:04 AM
Sorry, the way you listed, I missed the PCI-6281.
What were the results from my suggestions?
12-12-2007 09:34 AM