Signal Conditioning
Thermocouple measure in RSE mode using SCB-68
Dear Folks,
I am using PCI-6251 DAQ board and BNC-2090a and SCB-68 accesories for my experimental set-up.
Now I am trying to measure temp. using J-type thermocoupe and CJC feature in SCB-68.
I checked channel 0 for CJC and it shows 21 oC in RSE mode after change the scale according to the manual.
But if I use RSE mode in channel 1( I change the switich in BNC 2090a to SE) with J-type thermocouple, I could not get right temp. I wired Red to 67 pin (AI GND) and White to 33 (AI1) of J-type thermocoulple to SCB-68 terminal. I checked CJC enable box in SCB-68 configuration in MAX.  I also correctly set-up the jumpers in SCB-68.
What did I do wrong?
Help me....
(8,008 Views)
Suggest you configure the CJC for DIFF mode.  See this thread for solution to very similar problem.  http://forums.ni.com/ni/board/message?board.id=170&thread.id=377862
(8,003 Views)
Centerbolt is correct,
Your CJC Thermocouple needs a Differential Confguration.
See Knowledge Base below
How Do I Use the Cold Junction Compensation on the SCB-68?
http://digital.ni.com/public.nsf/allkb/96B996BAD82A84E9862570690066B03F
Juan Galindo
(7,983 Views)
I am facing the same problem,I am trying use CJC of BNC-2120 thermocouple Ic sensor for temperature measurement,but I am not able to acquire it with M6289 mass terminator device.
Whenever I make labview program using DAQ assistant,using ai1 as physical channel and ai0 as temp ref for CJC(default in device),It is giving daq error that task is not specified.
I don't know what is the program whether 6289 can control temperature sensor directly or not.
Please assist me in this regard.
(6,667 Views)
Hi Anuvohr,
Can you please tell if you’re getting a specific error code and post your code? Meanwhile be sure to review the information given in these knowledge bases: Reading Thermocouples Using CJC on the BNC-2120 and BNC-2120 Temperature Sensor on M Series Second Connector.
Have a good day!
(6,653 Views)