It seems like you are having incorrect software readings for these channels.
You may have a settling problem caused by either filters or high gains. First, check to see if you have enabled filtering. If you use a 4 Hz or 10 kHz filter on the SCXI-1100, you must scan channels at a slower rate. Because the filter circuitry is located after the amplifier, you must give the amplifier time to settle before switching to the next channel.
If you are not using the filter, check the amplifier gains you are using on the SCXI-1100 module and the DAQ board. Using a very high gain on your SCXI-1100 module and on some DAQ boards also has the effect of increasing your settling time. Decreasing your scan rate (increasing the delay between channels) should sol
ve this problem.
If these suggestions do not solve your problem, please reply to this post and I will look for a different solution to your dilemma. Have a great day!