Signal Conditioning

cancel
Showing results for 
Search instead for 
Did you mean: 

Use SCXI-1126 with other connector blocks?

I am short a connector block (SCXI-1320) connector block for a the SCXI-1126 freq module.  I have some extra SCXI-1300, 1302 & 1303 connector blocks.  Could one of these be used?
0 Kudos
Message 1 of 2
(5,707 Views)
 

Hi ranger,

Unfortunately, no, you can't use any of those terminal blocks with the 1126. The 1300 and 1303 terminal blocks are designed for 32-channel SCXI modules. The 1126 has 8 channels and so the screw terminals will not correspond to the correct channels. The 1302 can only be used with the SCXI 1180 feed-through module.

There are several terminal blocks that are compatible with your 1126. In addition to the 1320, you can also use the 1305, 1327, and 1338. If you're interested in any of these, please look at their product pages to see if they meet your needs.

Please let me know if you'd like further clarification.

 
Joe Friedchicken
NI Configuration Based Software
Get with your fellow OS users
[ Linux ] [ macOS ]
Principal Software Engineer :: Configuration Based Software
Senior Software Engineer :: Multifunction Instruments Applications Group (until May 2018)
Software Engineer :: Measurements RLP Group (until Mar 2014)
Applications Engineer :: High Speed Product Group (until Sep 2008)
0 Kudos
Message 2 of 2
(5,703 Views)