Signal Conditioning

cancel
Showing results for 
Search instead for 
Did you mean: 

What is the minimum Voltage that the SCXI-1104 can read?

I need to use a NI board to log voltage data from a system, and need to know what the minimum voltage the SCXI-1104 board can measure.  Could it perhaps go as low as 0.8 mV?

 

Thanks

0 Kudos
Message 1 of 2
(5,779 Views)

The input range for the SCXI-1104 is ±60 VDC/±42 VAC.  It is not made for mV measurements.  The manual states:

 

Can you use the SCXI-1104/C for mV measurements, such as temperature measurement using thermocouples?

No. The SCXI-1104/C is optimized for medium-voltage measurements, not low-voltage measurements. If you need to make low-voltage measurements select the SCXI-1102/B/C instead.

 

You might be able to get this resolution but it will not be very accurate at that low of a voltage.

Regards,
Jordan F
National Instruments
0 Kudos
Message 2 of 2
(5,771 Views)