01-22-2009 10:35 AM
I am monitoring the voltage output of a 48V DC power supply using an SCXI-1121 with SCXI-1327 attenuator. My SCXI chassis is configured in parallel mode, communicating through the backplane to a PXI-6040e. I have the hardware and software settings for all gains set to 1 and the attenuation switches / software set to 100.
When I view ai0 in the test panel in MAX, it is always at -5V (or whatever the bottom-end of the allowed range setting I have specified - even -300V) regardless of the power supply being on or off.
I checked the voltage directly at the screw terminals of the SCXI-1327 with a multimeter and it reads at 48V. Any ideas on what could be causing this issue?
I would appreciate any information you can provide.
Thank you,
Nick
01-26-2009 12:01 AM
Hi Nick,
Thank you for posting on the discussion forums.There's another thread that talks about a similar problem. PLease check it out at the following link.
Best Regards,
Faris A
01-30-2009 01:41 PM
Bueller,
Thank you for the response. It wasn't the same problem as in the other posting, but was rather a misunderstanding on my part about how parallel mode works. I essentially had no signals connected.
Regards,
Nick
02-04-2009 01:12 PM