Signal Conditioning
Working with SCXI-1143
I am working with multiple signals which have peak to peak voltages of 30 to 50V.
I want to filter out the noise above 20Hz for these signals.
I saw SCXI-1143 as a possible candidate. but voltage spec is low (max input is 5V)
Is there any attenuator module for this programmable LPF module.
Regards
(4,681 Views)
I want to know whether the SCXI configuration below will work or not.
Can some one help me on this.
The High Voltage signal (50Vac) will collected by an SCXI-1125 with a terminal module of SCXI-1313.
Output from back plane will go to SCXI-1143.
The filtered out put is acquired by PXI-6289 via a SCXI-1349 Adapter /SHC68 -68EPM  cable.
(All these modules should work in Parallel Mode but not a must)
Rg.
Naushad
(4,680 Views)
Hi Naushica,
You should be fine with that configuration the SCXI-1313 has a programable 100:1 attenuator so you will be able to acquire signals +-300V. There is a filter on the SCXI-1125 but it is limited to either 4Hz or 10KHz (or off).
Regards
Jon B
Applciations Engineer
NI UK & Ireland
Message Edited by JonBowers on 06-26-2007 05:58 AM
(4,666 Views)