04-11-2008 07:06 AM
04-14-2008 04:32 AM
Hi,
the TC-2095 is a connector block- which module are you connecting to? In principle, you should be able to use the same module as thermocouples also output mVolts. A few considerations:
1) Input range: make sure you don't exceed the input range of the module. Thermocouple modules usually have a 100x1 gain amplifier so you may need to disable that by setting your tasks accordingly.
2)CJC- Cold Junction Compensation must be programmatically disabled as it is only used for thermocouple measurements.
3) Thermocouple modules usually have a 2Hz low pass filter which you may also want to disable if you are interested in higher frequency responses.
Thanks,
KostasB
NI UK Applications Engineering
04-14-2008 02:07 PM