Signal Conditioning
change measurement range SCC-TC02
Hi!
I tryed to gauge temperatures with type-K thermocouples connected to SCC-TC02 modules on a SC-2345 board. The DAQ device ist an USB-6225.
My problem is that the upper limit of the measurement range is 153°C so temperatures above will be cut off.
Where can i change the range / gain of the amplifier?
Sorry for my bad english...
Thanks for your help
(6,322 Views)
Hey Kleinfeldt,
with the thermocouple type K you should be able to acquire a temperature more than 1000°.
Its important that you configure the SC2345 device as a assecory to the 6225 device.
http://www.ni.com/pdf/manuals/371064e.pdf (page 10)
But you can only choose the assecory on ta 68 pin device.
In the next step create a DAQmx Task for thermocouple and choose the correct input.
Kind regards,
Elmar
(6,309 Views)
Hi!
Thank you for your answer Elmar.
I created a DAQmx task and configured it.
Now i can measure temperatures above 1000°.
So thanks a lot!
Kind regards,
Ingo
(6,303 Views)