What you are using is not a strain gage - it is a load cell, so it doesn't make sense to try and scale it's output using one of our strain gage equations. I would recommend creating a DAQmx Custom Voltage with Excitation Task, because you have all the information that you need to scale your signal into kilograms based on the raw voltage reading. Let me explain...
At full scale your voltage reading will be 2.6304 mV / V, where V is refering to the excitation voltage input. So, if you use 10V for your excitation voltage, then at 5kg (which is full capacity) you will read 2.6304mV/V * 10V = 26.304mV.
Vex = Excitation Voltage (V)
Max Load = 5kg
Vr = Voltage Reading (V)
Now, since your load cell is very linear, you can compute your slope of the y = mx+b equation by saying that m = (max load / (Vex * 2.6304)), which yields the equation:
Load = (Max Load / (Vex * 2.6304)) * Vr * 1000 (the 1000 accounts for the conversion from your reading which is in Volts to mV)
or: Load = (1000 * 5 / 2.6304) * (Vr / Vex)
or: Load = 1900 * Vr / Vex
I hope this helps!