Hi,
I was trying do a custome volatage task for my HBM load cell (Type: C6A : capacity: 100 tonne). Is it necessary that I calibrate the instrument usiing the calibration wizard provided under the calibration tab inside MAX? My lab technical person told me that when I give the excitation voltage correctly and given my custom scale, the SCXi 1520 and 1314 (on a 1052 chasis and 6251 DAQ)should indicate the correct load. But this is not happening to my case. I have a few questions:
1. In the HBM data sheet for C6 force transducer, I searched for excitation volatage, it gives a range (say 0.5 V to 10 V). What wall I enter for the loadcell. I tried with 1 V , 2 V, ... 10 V. Each time when I click the calibrate wizard, I get different different values for known load values. But all the time the indicated load reading by NI is far from the actual value. I have applied the loads through an UTM and read the values using an HBM indicator. To check the HB indicator I also connected anothe indicator of different sort. Both showed the same reading. But when I connect it to the NI, it sensed different values: I will write the steps I followed to measure load value:
1. Create a custom voltage task in MAX.
2. Change Max: 100000 kg(f) (100 tonne capacity load cell: 2mV/V output: excitation 0.5 V to 10 V); Min to 0 kg(f)
3. Full bridge; Excitation: 1 V( I tried with 2, 3,4,5,6,7,8,9,10 but each time with different result).
4. Custom scale: Map range
prescaled: Maximum: 2 mV/V multiplied by 1 V excitation = 2m
Minimum: 0
Mapped value : Maximum = 100k ; Minimum: 0
Units: volts to kg(f)
5. Sampling mode: continuous sampling rate 1000; sampling frequency: 1000
5. Click Calibration under the third tab (after settings, device,)
6. I kept two load cells of 100 tonne capacity each of the same make one above the another between the cross heads of the UTM. One load cell was connected to an HBM load indicator and the other to the SCXI 1520 and 1314 terminal block channel 0.
7. No load applied; In the calibration table of DAQ task in MAX, the sensor showed zero.
5 tonne applied: sensor indicated 3.5 tonne in the calibration table (different 1.5 tonne)
10 tonne applied: sensor indicated 8.2 tonne. (difference 1.8 tonne)
15 tonne applied: sensor indicated ; sensor indicated 15.2 tonne
20 tonne: indicated 22 tonne:
25 tonne applied : indicated 27 tonne:
8. The above values showed dffrerent result when I changed the excitation voltage to 2 V and redo the calibration. I have done this upto 10 V excitation, each time it gives some absurd result.
I am totally fed up with these measurement system. Did I make any mistakes in the above process? I don't know. It would be of great help if somebody could help.
Regards,
Rajeevan