Before answering your original query,
something i would like to ask:
Do u want measure voltage on a DAQcard 6280's spare AI channel? or do you want to measure voltage on the SCXI 1521B's channel itself??
ps: both are possible, while measuring voltage on DAQ cards's AI channel is more straight forward, measuring voltage on SCXI 152, needs some settings to be changed
01-31-2008 09:36 AM
01-31-2008 05:50 PM
Hi Aaron,
The error you are seeing is occurring because you cannot use the same physical channel for different measurement tasks, such as strain and voltage. There are a few ways you can go about doing this. If you have another channel available and are using a full bridge configuration, you can add wires from the differential signal output of your strain gauge to an adjacent channel to do the voltage measurement, thus using two different physical channels. Another option, which you can use if you do not have additional analog inputs available or are using a configuration other than full bridge is to use a Custom Voltage with Excitation task. This will allow you to provide the necessary excitation for your strain gauge, but still read the raw voltage from it. The strain measurement can then be added by configuring this scaling in software manually. That is essentially what the strain task does under the hood. If you let me know what type of bridge you are using, I can give you further advice about the appropriate way to go about the measurement.02-01-2008 08:00 AM - edited 02-01-2008 08:02 AM