Signal Conditioning
open thermocouple detection
I am using an SCXI 1120 module with a 1320 terminal block to take temperature readings from k-type TCs. When we disconnect a TC the reading for that channel goes to around 272 C, which is in our measurement range. How can I get that reading to go out of range when the TC is removed?
(3,535 Views)
Greetings,
The following KnowledgeBase explains the concept of open-thermocouple detection:
Open-Thermocouple Detection
Since the SCXI-1120/SCXI-1320 combination does not have open-thermocouple detection circuitry, you will need to provide it yourself by connecting pull-up resistors between the CH+ terminals and voltage references. Since the channels of the SCXI-1120 are isolated, it is important to realize that connecting multiple pull-up resistors to the same voltage reference will break this isolation.
Good luck with your application.
Spencer S.
(3,535 Views)
Spencer, thanks for your response.
What is the importance of knowing my channels are non-isolated? (I don't know much about signal conditioning)
Thanks for your help,
Dave
(3,535 Views)
Dave,
If your input signals do not have differing common mode voltages, this will not be a big deal. However, biasing multiple channels to a common source will break the 250 Vrms channel-to-channel, common mode isolation.
Spencer S.
(3,535 Views)