Hi Flory
This 2.5V is the built in excitation voltage and cannot be changed for that device. The 2.5V is not the voltage that is output from the module. The link below should help you to configure the device.
http://www.ni.com/pdf/manuals/371073b.pdf
The configuration and Using the SCC-SGXX sections should be useful.
Let me know how you get on.
Regards
Beejal
11-27-2007 04:06 AM
11-29-2007 04:05 AM
Hi Flory
I have done some more research into you problem. The tutorial you were looking at is correct. However the SCC module that you are using has a built in excitation voltage of 2.5V and this cannot be changed through MAX.
I can suggest two options:
1) Connect a potential divider to the excitation voltage terminals of the module to reduce the voltage.
2) Connect an external power supply to your pressure sensor to provide the excitation voltage. Then use the SCC module to read the voltage output from the sensor.
Regards
Beejal
11-29-2007 09:21 AM
Hi Beejal,
Thanks for your suggestions! That was really helpful in clarifying the situation.
Regards,
Flory