Signal Conditioning
scxi 1520 exitation of 11 -30 vdc
Hello
I want to know wether i can use a device in my strain (scxi 1520) gauge module which requires an exitaion of 11-30 vdc and gives an ouptput of 4-20 mA.if possible whats the configuration and if not ,if i provide an external exitation how do i use it with my scxi 1520 module..
plz its urgent
lalit
(3,317 Views)
The SCXI-1520 has a maximum 10V excitation. Also, the 1520 is a voltage input module, it does not measure current. The SCXI-1121 will measure 4-20mA, but can also only provide 10V excitation.
-Alan A.
(3,309 Views)
You can convert almost any current output to a voltage output by adding a resistor across the output terminals. Calculate the resistor such that the voltage across the resistor at max current does not exceed max input voltage of the acquisition system. Also, check the specs for max output voltage of the signal source. Don't expect the circuitry to supply more than the circuit supply voltage minus, say, 20 p.c.
Of course you will have to compensate the offset - at 4mA 'live zero' output you will of course get a reading different from zero. This should be possible by software.
However, if you need an excitation voltage higher than supplied by the board you can use an external power supply. Maybe you will have to use a grounding scheme different from what be used with an 'internal excitation' source (from the board).
(3,282 Views)