Signal Conditioning
ventilation distribution in an SCXI chassis
I have a single SCXI-1102B module in a SCXI-1000 chassis. It is used for precise temperature measurements, where in particular the noise in the range 0.01 to 10Hz is required to be as low as possible.  Unfortunately, by some bureaucratic error, I got only one set of filler panels. There is no time to get the rest.
One easy way out is to cover the openings with blind panels cut to the size. This would probably strengthen the air flow through the one module.
Another way out is to cut one blind panel to cover the front opening and one perforated panel to cover the back opening, so as to maintain the ventilation through the single module to a level such as in the case of four modules being present.
However, it seems to me that the holes in the back of the modules are smaller not larger than the holes in the filler panel, so a filler panel would not equivalate the aerodynamic resistance of the holes in the back of a module + the inside circuit in the module.
So maybe the distribution of the flow of air between the modules and the un-occupied slots is not that important ?
The question is:
Is this matter of importance to the very low frequency noise level ? Or is this problem of a much lesser order of magnitude by it's influence and needs not to be addressed ?
(5,874 Views)
Hi Adrian,
You should be able to meet the noise requirements you have for the SCXI-1102B Module with or without the filler panels. The airflow for the module should be appropriate without the filler panels. If you are concerned about closing off the additional slots due to debris getting into them, the filler panels are a very good idea. The filler panels will improve the life and reliability of the chassis by ensuring that they are closed off from the environment.
Steve B
(5,854 Views)