I am trying to use SCXI 1102 to aquire analog signal(0~5 volts). Why I got 0 volts signal if the loop is closed(for example: If channel 10+ and channel 10- is connected together) while 5 volts signal if the loop is open(channel 10+ and channel 10- is not connected to each other)?
I am using SCXI 1000 chassis and SCXI 1102 connect to DAQ board(PCI-MIO-16XE-50). I can configure virtual channels for SCXI 1102. How can I configure virtual channels for the rest of my modules(I still have SCXI 1160, 1124, 1180)?
connector
(3,413 Views)
Amplifiers wothout any input load will "float" to a power supply limit, sometimes plus, sometimes -.You cam, not scan that channel, scan the channel and ignore, wire the channel to your field loop, or install a resistor across the terminals. I usually use a 10K resistor.
(3,413 Views)
Hello;
Instrumentation amplifiers amplify the voltage difference in between inputs - and +. In case you short the inputs together, the difference voltage will be zero, meaning you will get a 0V result.
In case you don't connect a valid signal source to the instrumentation amplifier inputs, its inputs will float, meaning the amplifier will saturate to its maximum reference voltage.
Hope this helps.
Filipe A.