Signal Generators
6733 AO slow rise time
I am using a PXI 6733 AO board as output on a closed-loop PID control. There appears to be excessive lag on the output when compared to the error signal, and as an experiment, I gave an unbuffered 0-10V step output command to the card. I measured the rise time for 0-10V at the output (no load, directly connected to SCXI 1121 for acquisition) at approximately 200mS. This is nowhere near the rated 2.8uS...am I doing something wrong?
(6,824 Views)
Hello Steve,
How exactly are you measuring the rise time?  You would need to be acquiring the voltage channel at at least 350 kS/s to register a rise time of 2.8 us.  How fast are you acquiring from the 1121?
Sean C.
(6,809 Views)
Sean,
I'm only acquiring at 1kHz, but I'm not really interested in the absolute value of the time, just need it faster than it is. My application is a real-time control loop operating at 500Hz, so 1k should be fast enough to show me the step response. In continuing my investigation, I have found that the 200ms time is constant regardless of the magnitude of step change (-10 to 10, 0-10 or even 0-1V). Makes me think that something somewhere is configured to generate a ramp, even though my code has set the buffer to zero. I also get the same rise time when I send a command from the test panel in MAX.
Steve
(6,802 Views)
Hi Steve,
Can you confirm that the rise time is 200 ms with an oscilloscope?
When acquiring at 1 kHz, does your input buffer show multiple points (around 200) transitioning from 0 to 10 Volts?
Sean C.
(6,790 Views)