02-01-2010 04:44 AM
Don't start multiple threads!
I didn't. I started one thread and then after searching on the subject, added a reply to someone elses thread who was obviously having similar trouble to me but on a different instrument supported by the same driver.
Also, It is the pnp driver, not IVI.
Thanks.
02-05-2010 05:37 AM
Found the problem.
When a MXG series Sig Gen does not have the modulate output option installed in the instrument and the labview pnp driver trys to set its state with OUTP:MOD OFF, an error is returned.
The driver needs to check if it the option is fitted before sending the command.