Signal Generators
NI-FGen and Amplitude Vp-p
A customer of ours sent us their PXI 5401 so we could write a C++ application with it. I put the 5401 in a chassis along with a 5102 and fired up the soft panels for the two devices. When I output a waveform, say 1 kHz with an amplitude (p-p) of 1V, it actually generates a 2V p-p signal (according to my eyes and the VirtualScope software ). I'm a software person that only knows enough about electronics to ruin them, so treat me kindly if I am mistaken. It would seem that the two devices would agree on the definition of V p-p.
Q. Do I have a basic misunderstanding of the definition of peak-to-peak?
Q. I can simply cut the desired amplitude P-P in half before sending to the NI-FGen API, but wouldn't I run the risk of only generating ha
lf the amplitude if NI decides to change their implementation in the future?
Q. The board has a metrology lab sticker on it from the customers cal. lab. Is it possible that their cal lab is the one with the inconsistent definition of Vp-p and calibrated the thing off by 100%? (that would seem pretty remarkable error correction).
Thanks for your patience and help,
Chris Lincoln
ULTRAX, Inc.
(7,125 Views)
What you see is the correct behavior from both hardware and software.
The NI 5401 creates the signal with the assumption of a load impedance equal to its output impedance of 50 ohm. The NI 5102 on the other hand as a input impedance of > 1Mohm.
When you use the Sources SFP you can select Edit>>Device Configuration and set the load impedance to > 1Mohm. The 5102 will then read the same voltage level you selected in the Source SFP.
If you program the 5401 from C++ it depends on the input impedance of the device/s you have your 5401 connected to. If your load impedance is effectively > 1Mohm, yes you can divide the amplitude by two before sending it to the NI-FGEN API (That is in fact what the Source SFP does).
Marian Koop
Software Engineer
N
ational Instruments
(7,125 Views)