Hi,
I have enable both the filters already. But I'm still seeing the same output.
Any idea why?
09-26-2007 10:29 AM
09-26-2007 10:44 AM
09-26-2007 11:07 AM
09-26-2007 11:56 AM
I tried to reproduce this on a PXI-5441 and a spectrum analyzer but I saw the exact same results in Function Generator and Arb Mode.
Are you getting underflow errors or are you padding your data somehow? Because you should not be able to generate a single 4-sample waveform in arb mode continuously at 100MS/s without underflowing (which would affect the ouptut signal considerably). Check to make sure that there is only one waveform in the "Waveform Array" in the Arb Waveform example and that you're not missing any errors. I added some labview code on my block diagram to download 4 periods of the 4 sample waveform - 16 points total - so it would not underflow.