Signal Generators
exporting mulitple marker events to different locations
I have a pxi 6561. If I have a script where I am exporting two marker events. How can I configure my device so that the two marker events are exported to different locations. I want to export marker event A to pxi_trig_0 and marker event B to pxi_trig_1.
Thank you,
-Tim
(7,880 Views)
Hi Tim,
Have you seen the example called "Dynamic Generation with Marker Event.vi"? Take a look at the vi called "niHSDIO Export Signal.vi" and notice how it is exporting that into a physical line.
The example is located under Hardware Input and Output » Modular Instruments » NI-HSDIO » Dynamic Generation » Scripted.
Hope this helps,
Gerardo O.
RF SW Engineering R&D
(7,866 Views)
Thank you for your reply. I had seen how to do this. I missed a connection on the vi.
Do you know if repeats can be nested inside of each other inside of scripts.
Suppose I have
repeat 9
repeat 10
end repeat
end repeat
as my script.
Thank you,
-Tim
(7,855 Views)
Hi Tim,
Yes you can do this. Take a look at this:
Nest repeat <N> or repeat until instructions inside a repeat forever instruction:
repeat forever
    <instructions>
    repeat <N>
        <instructions>
    end repeat
    <instructions>
end repeat
I found this on the NI Digital Waveform Generator under Start found in All Programs » National Instruments » NI-HSDIO » Documentation
Then you can browse for all scripting functions under: Programming » Reference » Scripting Instructions
Hope this helps,
Gerardo O.
(7,844 Views)
Hi Tim,
I noticed that the example I posted is not exactly what you were trying to do. I posted a nested repeat <N> inside a repeat forever. I must add then the note at the end of this same document I refer to:
Note  You can nest repeat <N> and repeat until instructions, one level deep, inside a repeat forever instruction. The if/else/end if instruction is allowed inside a repeat forever instruction, but if/else/end if is not allowed inside of repeat <N> and repeat until instructions. Other nesting (for example, repeat <N> inside another repeat <N>) is not allowed.
This means that the exact script you wrote as example is not allowed.
Let us know if you have other questions,
Gerardo O.
(7,838 Views)
Thank you for your message. That helps a lot.
-Tim
(7,830 Views)