Turning off IEPE in Signal Express with 9234?

Solved!
Go to solution

I have a 9234 module and I can't figure out how to turn off IEPE while using Signal Express.  A new tab named "Device" shows up when looking at 9234 modules, but that only has a pull-down list for choosing AC/DC coupling.  MAX shows a checkbox for IEPE, but I can't find that in Signal Express. 

 

Any suggestions?

 

Thanks

0 Kudos
Message 1 of 4
(7,287 Views)
Solution
Accepted by topic author ChrisCW

Hi ChrisCW,

 

In order to be able to set the IEPE you need to choose a nonvoltage task such as sound pressure or acceleration.  You will see in the picture that I have both IEPE source and value available.  If a voltage task is the only one that makes sense for your application you could still use one of the others and then scale the data.

 

I will go ahead and submit a product suggestion to have this added to the voltage task also.

 

Message Edited by Jason_D on 10-14-2008 09:29 AM
Sincerely,
Jason Daming
Applications Engineer
National Instruments
http://www.ni.com/support
0 Kudos
Message 2 of 4
(7,265 Views)

Doh!  I didn't even think of the "Iex" items.  I was looking for checkbox like in MAX.  So I assume setting "Iex Source" to "None" turns off IEPE?

 

Yeah, we are more likely to disable IEPE in the voltage task, i.e. inductive tachometer.

 

Thanks,

Chris

 

0 Kudos
Message 3 of 4
(7,246 Views)

You are correct, none should disable IEPE.

Sincerely,
Jason Daming
Applications Engineer
National Instruments
http://www.ni.com/support
0 Kudos
Message 4 of 4
(7,233 Views)