I have a 9234 module and I can't figure out how to turn off IEPE while using Signal Express. A new tab named "Device" shows up when looking at 9234 modules, but that only has a pull-down list for choosing AC/DC coupling. MAX shows a checkbox for IEPE, but I can't find that in Signal Express.
Any suggestions?
Thanks
Solved! Go to Solution.
10-14-2008 09:26 AM - edited 10-14-2008 09:29 AM
Hi ChrisCW,
In order to be able to set the IEPE you need to choose a nonvoltage task such as sound pressure or acceleration. You will see in the picture that I have both IEPE source and value available. If a voltage task is the only one that makes sense for your application you could still use one of the others and then scale the data.
I will go ahead and submit a product suggestion to have this added to the voltage task also.
10-15-2008 08:28 AM
Doh! I didn't even think of the "Iex" items. I was looking for checkbox like in MAX. So I assume setting "Iex Source" to "None" turns off IEPE?
Yeah, we are more likely to disable IEPE in the voltage task, i.e. inductive tachometer.
Thanks,
Chris
10-16-2008 10:35 AM
You are correct, none should disable IEPE.