Spirometer: Characterization of Lung Function

Spirometer:

Characterization of Lung Function

Rebecca Rabban, Joanna Salacka, Roy Zhang

INTRODUCTION

The project objective was to build a spirometer to successfully characterize lung function. A spirometer measures

volumetric airflow in and out of lungs and allows the diagnosis of lung diseases, such as asthma, which currently

affects 300 million people worldwide. Potentially fatal respiratory diseases, such as chronic obstructive

pulmonary disease and cystic fibrosis, can be detected and their treatment responses tracked. Additionally,

spirometer assists respiratory exercises, which are especially important after surgery or prolonged bed rest to

prevent lung collapse and pneumonia.

The device consists of a tube with two different cross-sectional areas, with the user exhaling into one end of the

tube. The design is based on the Venturi effect, which describes how volumetric air flowing through a tube

remains constant, air velocity increases as it moves into the tube with smaller cross-sectional area to create a

pressure difference. This differential is converted to volumetric flow rate by LabVIEW using the Bernoulli’s

principle. Other parameters to be analyzed include Peak Expiratory Flow (PEF), Forced Vital Capacity (FVC),

and Forced Expiratory Volume in 1 Second (FEV

1). PEF, in liters per second, is the maximum flow rate achieved

by forced exhalation after a full inhalation. FVC, in liters, is the maximum volume forcibly expelled after full

inspiration. The third parameter, FEV

1, is the maximum volume of air that can be forcibly blown in the first

second.

EXPERIMENTAL SETUP

Our system is composed of a flow transducing device (Venturi tube), differential pressure sensor, conditioning

circuits and LabVIEW VI. The pressure sensor measures the difference in pressure between sections of the

Venturi device and outputs a voltage proportional to this difference, in the 20 mV-50mV range. The conditioning

circuits consist of a differential amplifier with a gain of 100 and a low-pass filter with a 3dB rolloff at 26 Hz.

These circuits serve to increase the signal magnitude to the volt scale and filter out high-frequency noise

(especially 60 Hz noise). The frequency spectrum of interest for spirometry is typically 0-20 Hz.

We calibrated our sensor and signal conditioning circuits as a single system in order to obtain an equation for

conversion of voltage output to measured pressure difference. We accomplished this by applying a range of

known pressures to the sensor, using a pump and manometer from a blood pressure cuff, and recording the circuit

output.

The conditioned signal is acquired and analyzed by our LabVIEW VI. The data acquisition portion of the VI reads

the voltage data and converts the values to pressure difference using our calibration equation and then converts

the pressure difference to volumetric flow rate (Q) using the theoretical relationship above. Once a measurement

is complete, the analysis portion of the VI integrates flow for the entire data set to obtain volume vs. time and

flow vs. volume. It also calculates the parameters PEF, FVC, FEV

1 and FEV1/ FVC. The VI front panel displays

these parameters as well as graphs of the volume data.

RESULTS

Appendix A shows an example of the waveform results from the front panel of our VI. This specific result is for

our “anaerobic athlete” test subject. Table 1 summarizes our results for each test subject.

We were able to obtain valid physiological values for PEF, FVC, FEV

1, and the ratio between FEV1 and FVC.

We also saw expected trends between subject types with 2 exceptions. The obstructive disease subject’s FVC was

greater than that of our average male subject, although we had expected it to decrease. This is most likely because

the subject was seven inches taller than the average subject. We had also expected the smoker to have a lower

PEF than average, but his peak was actually greater. We are unsure of the reason for this; it is possible that his

lungs are not severely affected by smoking, since he began smoking recently.

BLAH.bmp

Each subject performed a minimum of 3 trials and the device was able to produce similar results for each trial. In

addition we are able to obtain consistent results over multiple days. We consider the device an acceptable proof of

concept because it obtains expected physiological values and can differentiate between a normal and abnormal

lung.

DISCUSSION AND CONCLUSION

Our project was to build and test a proof-of-concept prototype of a low-cost spirometer. We have concluded that

our system is an acceptable proof-of-concept; although additional improvements and tests would be desirable, we

have shown that such a system is indeed feasible.

The primary source of error in our system was the lack of calibration of the Venturi device itself. We are

confident in our calibration of the electronic sensor and circuitry, but for full accuracy the flow transducing device

should also be calibrated with a known flow source to determine if and how the conversion from flow rate to

pressure difference varies from the theoretical equations. If this effect turned out to be significant, some

adjustments to the device would be needed in order to bring it closer to ideal behavior. Possible parameters to be

adjusted include: tube lengths, location of pressure sensing points, length of hose connecting these points to the

sensor port and funnel design. We also discovered that the sensor output varied with supply voltage, which could

affect our system calibration, so a DC voltage regulator would be useful for powering the differential pressure

sensor.

APPLICATION

A functionality that could be added is a disposable or reusable mouth piece, which grants health care

professionals added assurance of patient health and safety by minimizing disease transmission. This can be done

by covering the inlet opening on the spirometer with an easily replaceable piece of tubing. Additional

functionality added to the system would be having ability in LabVIEW to email spirometer test results directly to

health care providers, who analyze the results and record them for disease or recovery progress tracking. This

increases productivity and sources of revenue by allowing for greater patient outreach. In addition, patients

benefit by eliminating the need to travel to a health clinic to receive care. Also, it saves time and effort being able

to send test results directly instead of fumbling with an extra email client. This functionality can be implemented

by placing a SMTP Email Send Message VI on the LabVIEW block diagram, which would have a case that when

executed, sends email with pertinent data to a list of email recipients at the user’s discretion.

Attachment: Original Report