Switch Hardware and Software
How to use "scanner advanced" signal with DMM 4060?
Hi!
We're using SCXI-1128 switches with 1331 terminal blocks. The switches should communicate with the PCI DMM 4060 via the SH9MD cables.
Does someone know if there is a possibility to use the scanner advanced signal from the switch over these cables (for handshaking)? The documantation says the signal is only available on the front connector (1331) which does not make much sense in my eyes...
I'd be glad if someone could help me here.
Jan
(7,679 Views)
CHeck with NI sales on this one.
If I remeber correctly, you need a special cable to do what you wnat.
Ben
(7,678 Views)
The scanner advanced signal is available on both the front and rear connector of the SCXI-1128, however the SH9MD overloads the use of the scanner advanced pin on that cable to also control the SCXI chassis. If you want to do full handshaking through the SH9MD, you need a separate chassis controller such as the PXI/PCI-4021.
(7,678 Views)
When you are using the SH9MD to communicate from the 4060 to an 1128, you are using the cable for SCXI communication. You are connecting this cable to the back of the 1128 and the only signals it can carry are SCXI communication and the Measurement Complete (MC) from the DMM to the Switch. The 1128 cannot route his Scanner Advanced (SA) signal to the back where the SH9MD connects, and there are not any available PIN to carry it.
If you want to use handshaking with a PCI 4060 and a 1128, you will need to control the Switch using a different controller than your PCI 4060 (another 4060/4070/4021) connected to the back of the 1128 using the SH9MD cable. You will also need to get the "trigger cable" for your PCI 4060 (available page 298 in 2002 NI catalog). This cab
le splits the SH9MD into 2 BNCs; EXT TRIG IN and MC. These BNCs need to be connected to the SA screw terminal and to the EXTRIGIN screw terminal respectivelly on the 1331 terminal block.
The DMM will need to be set to external for both the Trigger/Sample trigger and the MC destination. The Switch would also have its SA destination and input trigger set to "front of module X".
A very good source of information in the "NI digital multimeter help" available on the Web. Check out the Triggering section under the Programming heading.
Good luck.
(7,679 Views)