05-27-2016 05:15 PM
In the user manual the maximum operate time (page 3) is listed as 16 ms. That's a typo right? It should be 16us right?
Solved! Go to Solution.
05-31-2016 04:03 PM
Hi AbandoningCausality,
It appears that this is most likely a typo. For reference, the 2536 uses the same hardware as the 2535--it just has a different topology. So, the maximum operate time should be the same 16us reported for the 2535.
I've gone ahead and filed a Corrective Action Request in order to update the spec sheet. This request can be tracked with the number 589816. Thank you for bringing that to my attention!