First time using a FET switch card, running basic vi where I connect a row and two columns and do a resistance measurement with a DMM.
I found that with the DMM connected, ie signal applied, the switch can take more than 5 seconds to close. (DMM ohm reading 26 ohms after 5 seconds, although the 'wait for debounce' vi reported all good)
If the DMM is not connected, it happens instantaniously.(disconnect DMM, perform switch action, connect DMM - instant closing)
Anyone experienced the same ?
Solved! Go to Solution.
01-13-2016 07:06 AM
Found the problem.
Seems the FET card goes into protection mode if you dont connect the DMM - signal to chassis GND.
After doing that, it worked fine.