I send a SYSRESET to a GPIB-VXI/C. Then, when i try to communicate (visa read message) with other instrument in the VXI, i get the -1073807265 error (NRFD & NDAC -> no listners).
To resume communication with instruments, i need to close all visa sessions.
My program runs under LabVIEW 6.1 (usually no need to open/close visa sessions). However, i observed same behaviour of my system when i open visa sessions (one per instrument and one for the controller) under MeAsurementeXplorer and send the SYSRESET command.
I found the following note in the knowledgebase reporting the same kind of problem, but not solving my problem:
After sending the SYSRESET command to
a GPIB-VXI/C, all subsequent commands return ENOL (No listeners found).
Hardware: VXI/VME/MXI>>VXI Interfaces>>GPIB-VXI/C
Problem: After sending the SYSRESET command to a GPIB-VXI/C, all subsequent commands return ENOL. This is happening because the GPIB-VXI/C needs some time to rerun RESMAN after the VXI reset. Wait until the sysfail light goes off and then a few more moments (after the sysfail goes off, there is some time needed for RESMAN to run) and then try a GPIB command (e.g. ibwrt "*IDN?") to the GPIB-VXI/C.
Solution:
Also, check for the GPIB-VXI/C configuration. It might be possible that the wait period (wait for sysfail to be removed from a vxi backplane) is larger than normal values (5 sec...)
Related Links:
Attachments:
Report Date: 25/11/1996
Last Updated: 11/08/1997
Document ID: 0ROGKM2Q