Your issue might be resolved by this KB article:
Error -1074388479 With the NI-9853 Module for CompactRIO
This specifies the CompactRIO and the 9853, but you should follow the same instructions.
If this does not resolve it, does your RT system have the same version of the CAN driver as the Windows development computer? You can check this in MAX.
If it has the same version, when you have gotten this approach to work in the past, was it on this same RT system and development computer?
12-08-2009 09:20 AM
12-08-2009 09:30 AM
NI PXI 8464 has as default different transceiver settings (Highspeed and LowSpeed)
For port to port tests you need to change them to identical settings
Hope that helps.
12-08-2009 09:43 AM
12-08-2009 09:44 AM
How do I change them on the RT target and what would I change them to?
-Eric
12-08-2009 09:59 AM
you can change the default e.g. in MAX >> Utilities >> NI-CAN >> RT Hardware Configuration
12-08-2009 10:16 AM
Wow, I can believe I didn't remember to do that. That got it working.