Dave,
Hi, you brought up several questions / issues -- let me see if I can help with some of them:
I can set this up CTR0 with the Count pulse (and possibly add the start pulse as an Arm Start....I can't use a Start Trigger, right?).
Yes, you could set up this way. Also, as far as I know you're also correct that you need to configure for an "Arm Start" trigger using the DAQmx Trigger property node. The "Arm Start" trigger is used for counter input (measurement) apps while the regular "Start Trigger" can be used for counter output (pulse generation) tasks. I don't think I've experimented with recent versions of DAQmx though so it may have changed in 7.4 or 7.5
...I tried to set up CTR1 along the lines of Gen Dig Pulse Train-Finite-Dig Start.vi for generating a finite pulse train starting on a digital trigger, and connecting the output from CTR1 to the Sample Clock Source on CTR0. However, I get an error -50103 saying the specified resource is reserved if I do both at the same time
I highlighted the problem -- the FINITE pulse train. DAQmx uses CTR0 as a "helper" when you generate a finite pulse train on CTR1. It would generate a single pulse whose width corresponds to the exact amount of time needed for CTR1 to generate its specified # of pulses.
For your specific app, I think you could generate a triggered continuous pulse train with CTR1 -- this wouldn't need to use CTR0 as a helper. The Start Pulse would arm CTR0 at the same instant that CTR1 is started. If you set up CTR0 to acquire on the trailing edge of CTR1's pulses, then you'll get the time bins you want.
A final slight mod would be to setup CTR0 for measuring buffered periods (set units == "Ticks") instead of counting edges. In that mode, you wouldn't have to do the subtraction at the end.
This is an unrelated topic, but is there a discussion of the relationship between the terminology in the manuals describing the cards (SOURCE, GATE, OUTPUT terminals) and the terminology in NidaqMX (Source Clock/SrcClk.Source, CI.CountEdges.Term, CO.Pulse.Term)? Eg. is CTR0.GATE always the same thing as SrcClk.Source, or does it depend on the mode of operation?
There's an NI app note and some discussion forum hits if you search the site for "daqmx terminology."
If the M-series hardware manual says to connect something to the SOURCE input, how do I assign an alternate PFI pin to that SOURCE input in Labview? Does it depend on the counter type, or is it always the same?
Usually, that choice would be available under the DAQmx Channel property node. There'll be some place to define where the input signal is coming from, generally with "Term" or "Terminal" as part of its name. Sorry I can't be more specific as I'm not at my LV computer now.
Happy counting!
-Kevin P.