Hello,

I´d like to measure the postion from a Haidenhein 2571 TTL linear encoder with labview 8.5.
http://wwwpdb.heidenhain.de/ansicht/download.php?FilePath=..%2Fansicht%2FHeidenhain%2Fmedia%2Fimg%2F&FileName=MA_MT_2571_ID33526592.pdf

The encoder is connected to a c9401 card in slot6 of an c9172 rack, all other slots are filled with analog measurement cards.

The hardware setup seems to be correct, since a get a correct signal with buffered and unbuffered aquisition of the encoder data (i.e. samples from the labview samples library.

The problem is: This works only for rather slow movements of the linear encoder. With higher travel speeds a loss of counts occurs.
The maximum speed without losses is about 65000 counts per second.

I have tested this with an buffered aquisition using the cdaq1/aisampleclock whilst running an analog read task with the analog module in slot1 with sample speeds from 10-100khz and also with unbuffered aquisition.

Could this be caused by bad cabling from the encoder electronics to the 9401 card or is it more likely a problem with my labview Vi?

Thanks for your answers

Ingo
0 Kudos
Message 1 of 7
(4,726 Views)
What are the interpolation (x5 or x10) and the scanning frequency (25, 50, 100 or 200kHz) of the integrated digitizing and interpolation electronics ? Do you have a scope to analize the output signals of the MT 2571 ?
0 Kudos
Message 2 of 7
(4,719 Views)
The encoder specifications are: TTLx5 , 100khz, which lead to vmax of 200mm/second according to its datasheet. I don´t believe i reached such high speeds during testing.

Do I have to adjust my labview VI to these specs in some way?

I may have access to a scope if the problem won´t solve but i also have another linear encoder for testing.

Ingo
0 Kudos
Message 3 of 7
(4,718 Views)
I would first check the hardware with a scope. The signals may be noisy, the levels may be wrong, the max output frequency of the length gauge may be reached... Do you also use the inverted signals ?
0 Kudos
Message 4 of 7
(4,714 Views)
I don´t use the inverted signal yet. How can i do that?
I can only find the following connection plan in the DAQmx help:
 
A: PFI 0
B: PFI 2
Z: PFI 1
 
ingo
0 Kudos
Message 5 of 7
(4,710 Views)
This needs an external differential line receiver like the 26LS32. The usage of the inverted signals is recommended in case of noisy environment or long cables.
You will find a schematics on page 41 in HeidenHain's length gauge user manual.
0 Kudos
Message 6 of 7
(4,697 Views)

Thanks for your support!

I will consider using such a differential line receiver if I encounter problems in the actual test setup.

But I hope the maximum speed will be enough for my measurement.

0 Kudos
Message 7 of 7
(4,694 Views)