I'm using the Measurement Computing USB-1608G which says it can handle 250KS/s per channel. However in DasyLab, I'm having trouble collecting analog samples at a rate that high.
I'm trying to collect data at that rate on 10 channels. Through the settings, I've found out how to set it to 200KS/s, but once I hit the PLAY button, I'm told that rate is too high for that many channels. I'm told I have to bump it down close to 20K/s which is far too slow because I'm trying to capture events occuring within 10's-100's of microseconds.
Is this a limitation of DasyLab? Is there a different setting I need to configure?
08-18-2014 04:55 AM
No, not per channel, 250kS/s is the USB-1608G's combined/sum/aggregate rate for all channels.
One channel >>> max 250kS/s
Two channels >>> max 125kS/s per channel
etc
http://www.mccdaq.com/PDFs/manuals/USB-1608G-Series.pdf (page 11)
08-18-2014 07:22 AM
Measurement Computing wrote the driver, and it is aware of the features and limitations of the supported products.
The USB-1608G uses successive approximation, and has one A/D converter for the channels. As Herr Wons has said, the maximum sample rate is aggregate, and you need to divide by the number of channels.
You should be using the USB-1616HS (1 million s/sec) or perhaps the USB-1608FS-Plus.
USB-1608FS-plus
See this product comparison
http://www.mccdaq.com/products/USB-1608-Comparison.aspx