DASYLab
Pressure curve sampling
I need to sample a pressure curve from a flame passing down a pipe the signal is 0 - 10V scaled to represent the pressure sampled by a Iotech DAQ3000.
The standard I am working to states the following
"The average value of the detonation pressure shall be calculated from the area integral below the pressure-time trace, starting at maximum pressure peak and covering a time interval of 200 µs."
I need to trigger the sample at the maximum pressure peak then continue for 200 µs and make the calculation stated.
Message 1 of 12
(8,631 Views)
Try the Min/Max module, configured for TTL pulse, and then use the TTL as the input to a Combi Trigger, configured for the post trigger settings that you need.
Use the output of the Trigger to control a relay where the input is the pressure signal, and you'll get a snapshot of the signal where you can then run the Integration.
Measurement Computing (MCC) has free technical support. Visit www.mccdaq.com and click on the "Support" tab for all support options, including DASYLab.
Message 2 of 12
(8,630 Views)
I am using Dasylab V9 and there doesn't seem to be a TTL Pulse only output modes are Function Value and TTL Peak. Using TTL Peak just gives me a list of 0 or 5 could you give me a worked example showing the settings for each module?
Message 3 of 12
(8,628 Views)
Look under Statistics for the Min/Max module.
Message 4 of 12
(8,627 Views)
I found the min/max module but there isn't a TTL pulse option only TTL Peak
Message 5 of 12
(8,625 Views)
That's the setting. I don't have V9 available, and I picked that out of my failing memory. Sorry.
The output of Min/Max is either a value or a 0/5 signal, where it outputs 5 at the time of the peak, and 0 for the rest of the signal.
Message 6 of 12
(8,623 Views)
Here's my V11 worksheet
22907iDEC971C2AE3D37C6
and the module settings. Note that for a high data rate, I had to turn off the "Limit data flow (FIFO size)" settings, making it possible that DASYLab will consume lots and lots of memory.I'm not sure whether V9 had this feature or always limits.
22909iD83280A14F20C69C
22911i7D4FF563BA3D29DA
22913iA78EF33A04EA6047
Message 7 of 12
(8,622 Views)
I am getting an error message Relay03 mismatched input channels.
Message 8 of 12
(8,615 Views)
How did you configure the Min/Max? It should be the TTL mode, not the value mode.
Message 9 of 12
(8,608 Views)
Apologies after some tweeking of the settings that were not detailed I think your solution will probably work for me.
Message 10 of 12
(8,605 Views)