Hi, we used PXI-4462 for data acquisition when measuring sound pressure. It worked good.
But for make our device more portable, we bought USB-4432, and want to use it instead of PXI-4462. The problem now is that the data obtained through USB-4432 have high level background noise, which didn't happen before.
The thing is, we used the same program for PXI-4462 and USB-4432, the data from PXI are good, but Data from USB have high level noise. Is there any special set needed for USB-4432? or the data acquisition qualities are different between PXI and USB?
07-06-2009
11:49 AM
- last edited on
02-15-2024
03:27 PM
by
migration-bot
Linda,
The accuracy of these cards are quite different as is also reflected in the price. The 4462 and the 4432 have many specifications that are different. I would suggest that you consult the detailed specifications located at each of these links. This very likely explains the behaviour you are seeing. This is a very useful article on obtaining the accuracy of each device.