Digital I/O
DC voltage measurement of a series coil circuit with NI 9205
Hi all,
I need to measure the DC voltage across a circuit comprising 24 coils in series. The voltage across each of the 24 coils ranges from 0 to 10 mV DC (much less than 1 V) . I was looking to use an NI 9205 module (on +/- 200 mV range).
Would the NI 9205 be suitable for this application and what connection arrangement is recommended.
I was thinking that I could not use differential measurement of voltage because of the series coil circuit as I would need to connect say the AI Ch0- of Coil 1 to the AI Ch1+ of Coil 2 etc. Or does the NI 9205 multiplexer negate this concern?
The other option would be to use the nonreferenced single-ended (NRSE), connect a common section within the series coil circuit to AI Sense terminal and then connect AI Ch0+ to Coil 1 and AI Ch1+ to Coil 2 etc. I would then have to calculate the difference in voltage in Labview to determine the individual voltages across the coils.
I would appreciate any comments or advice on the suitablity of the module and recommended wiring connections.
Thanks for your help!
Peter.
(2,726 Views)
Series coil circuit for information.Series coil circuitSeries coil circuit
(2,714 Views)
I believe you should be able to use the NRSE configuration, and then just programmatically calculate the individual voltages in LabVIEW. Ensure that you use to th e+/-200mV range so you can detect that the 10 mV.
Brian.D
NI Product Owner
VeriStand
(2,679 Views)