Digital I/O

cancel
Showing results for 
Search instead for 
Did you mean: 

USB DAQ 6001 Resolution

In USB DAQ 6001, when collected data using analog input, 14 mV signal is coming without any actual input.  

 

 

How I can correct this issue?NI.jpg

0 Kudos
Message 1 of 3
(1,125 Views)

The USB-6001 is a low-cost, multifunction DAQ device. It offers analogue I/O, digital I/O, and a 32-bit counter. The USB-6001 provides basic functionality for applications such as simple data logging, portable measurements, and academic lab experiments.

0 Kudos
Message 2 of 3
(1,108 Views)
  1. For any data acquisition device, inputs should not be left floating as the charge could build up due to the high input impedance (no worries if you will not be reading the signal)
  2. Short the analog input to AGND and see if you observe the same
  3. Please note that the typical accuracy is 6mV and worst case is 26mV
    santo_13_0-1643990805738.png

     

 

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 3 of 3
(1,101 Views)